Once Annealed Highly Cross-Linked Polyethylene Exhibits Low Wear at 9 to 15 Years
Orthopedics Posted April 18, 2016 DOI: 10.3928/01477447-20160414-03
James D'Antonio, MD; William N. Capello, MD; Rama Ramakrishnan, MS


Abstract

A once annealed highly cross-linked polyethylene (HXLPE) was introduced in 1998. Concerns regarding its long-term performance and oxidative resistance exist because of the presence of retained free radicals. The authors studied 48 patients with 50 hip implants having an average age of 62 years. They were followed for 9 to 15 years. The purpose of this study was to determine linear wear rate and the incidence of osteolysis and/or mechanical failure. At an average follow-up of 12.2 years, the annual linear wear rate was 0.018 mm. No mechanical failures or osteolysis have been found to date. The clinical performance of this HXLPE continues to meet expectations despite the presence of free radicals.

HXLPE was introduced in 1998 to reduce wear and osteolysis experienced with conventional polyethylene. Cross-linking is accomplished by first irradiating the material with gamma or electron beam irradiation (5 to 10 mrad) and then thermal processing the material either above (re-melting) or just below (annealing) the melting point. Although the former quenches residual free radicals at the expense of an initial reduction in physical properties, the latter maintains the physical integrity of conventional polyethylene while leaving free radicals within the processed material.

Since their introduction, annealed and re-melted materials have shown similar and dramatically reduced polyethylene wear rates and osteolysis when compared with conventional materials.4–8 However, concerns remain regarding the mechanical properties and long-term wear characteristics of these first-generation materials. There have been isolated reports of fracture of re-melted HXLPE but not of the annealed material. Several studies comparing once annealed with re-melted HXLPE using accelerated artificial aging found oxidation, decreased ultimate strength, and higher wear with the former. Contrary to these artificial aging predictions, the clinical performance of once annealed HXLPE out to 10-year follow-up has continued to show low wear without osteolysis or mechanical failure.

Even with these positive reports to date, concerns about the long-term performance of once annealed material continue because of the presence of free radicals. The fundamental question and concern is whether the presence of any amount of free radicals in an HXLPE will eventually result in component failure through locking mechanism disruption or accelerated wear and subsequent osteolysis.

The objectives of this study were to
(1) determine the wear rate of once annealed HXLPE out to 15 years;
(2) report on an analysis of plain radiographs for evidence of wear debris–related osteolysis; and
(3) report any mechanical failures, including locking mechanism issues.​

Materials and Methods
A group of 64 nonconsecutive patients underwent 66 primary total hip arthroplasties performed by one of the authors (J.D.) between January 1999 and May 2000 and were prospectively followed. Patients were included if they provided written informed consent, were younger than 75 years, and had non-inflammatory pathology. During this same time, primary total hip arthroplasties with other bearing surfaces were performed for 97 patients (104 hips); however, these are not part of this study. Of the 64 patients, 5 (7.8%) died, 6 (9.3%) were lost to follow-up, 4 (6%) had clinical data from mailings but no radiographs after 5 years, and 1 (1.5%) underwent revision for recurrent dislocations. Clinical data included Harris Hip Scores15 and the rate of adverse events. Radiographs were analyzed at 6 to 7 weeks, 1 year, and 2 years postoperatively and at last follow-up out to 15 years.

Forty-eight patients (50 hips) with an average follow-up of 12.2 years (range, 9 to 15 years) had complete clinical data. Twenty-two of the 48 patients (46%) were male with an average age of 61.8 years. The average age of the females was 62.3 years at the time of surgery. The age range for all of the patients was 36 to 74 years. Osteoarthritis was the most common diagnosis (86%). Mean weight was 196lb (range, 110–340 lb). Four cases had poor-quality radiographs at final follow-up, leaving 46 cases with good-quality radiographs for wear measurement. Good-quality radiographs have sharp contrast between the acetabular cup and the pelvic bone as well as between the cup and the femoral head, permitting wear analysis software to identify the edges precisely for accurate wear measurement and cup position.

Once annealed HXLPE (Crossfire; Stryker Orthopaedics, Kalamazoo, Michigan) was manufactured from ram-extruded GUR 1050 ultra-high-molecular-weight polyethylene that had been irradiated to 7.5 mrad and annealed for 8 hours at 130°C. After machining, the Crossfire inserts were then sterilized with 3 mrad of radiation in nitrogen.

All patients received the same tapered titanium cementless femoral stem (HA Omnifit Hip Stem; Stryker Orthopaedics) and cementless titanium hemispherical acetabular shell (Secur-Fit; Stryker Orthopaedics) with a 28-mm cobalt chrome head and a 0° Crossfire liner.

Clinical data to determine the Harris Hip Score were collected at 6 to 7 weeks, 1 year, 2 years, and varying times throughout 15 years. At each visit, the authors also obtained an anteroposterior radiograph of the pelvis centered over the symphysis pubis, an anteroposterior radiograph of the hip centered over the femoral head, and a table-down lateral radiograph of the hip. The patient was in the supine position for all views. The authors digitized the anteroposterior radiographs of the pelvis with a high-resolution optical scanner (Vidar Corp, Herndon, Virginia) and used Martell software (Hip Analysis Suite; University of Chicago, Chicago, Illinois) to analyze linear head penetration from baseline (6 to 7 weeks) to 1 year, 2 years, and last follow-up postoperatively, cup inclination, and anteversion. All radiographic analyses were performed by one of the authors (R.R.). The 1-year radiographs were used to estimate initial bedding-in rates. Wear from 1 year to last evaluation then served as the basis for the yearly linear and volumetric wear rates. Mean volumetric wear rate (V) was calculated using the geometric formula

V=Pi*R*R*W

where Pi (the ratio of the circumference of a circle to the diameter) is 3.1416, R is the radius of the femoral head in millimeters, and W is the mean 2-dimensional wear rate. The formula is based on a cylindrical wear pattern perpendicular to the face of the cup.

Hips were evaluated radiographically for osteolysis in acetabular component zones as described by DeLee and Charnley and in all femoral Gruen zones on anteroposterior and mediolateral radiographs by 2 of the authors (J.D., W.N.C.). Implant stability was evaluated according to criteria described by Engh et al. The distribution of wear data was examined with the goodness of fit test. The wear data are presented as mean and SD. SAS statistical software version 9.1 (SAS Institute Inc, Cary, North Carolina) was used for data analysis.

Results
The wear rate at recent follow-up, excluding negative values, was found to be gamma distributed [goodness of fit test, P=.158, scale (sigma)=0.012, and shape (alpha)=2.01]. The mean linear wear rate was calculated as 0.018 mm/y (SD, 0.024 mm/y) at an average of 12.2 years. No patient had a wear rate greater than 0.085 mm/y, and this rate occurred in a patient who was 45 years old and weighed 265 lb (body mass index, 35 kg/m2) at the time of surgery. The wear data were sorted and plotted as a bar chart. Volumetric wear was calculated to be 11.1 mm3/y.

The 1 revision (1.5%) in the study was for recurrent dislocations. There were no fractures of the bearing surface, and no revisions are impending. The mean Harris Hip Score at last follow-up was 94.4.

All acetabular and femoral implants are well fixed. The mean cup inclination was 50.2° (SD, 5.1°) and the mean anteversion was 18.6° (SD, 8.3°). One patient was thought to have an early osteolytic lesion in Gruen zone 7 at 5 years. However, recent radiographs (at 15 years) showed no change or slight regression and it is no longer considered an osteolytic lesion. No other patient had evidence of osteolysis on either the acetabular or the femoral side.

Discussion
The purpose of this study was to report on the linear wear, incidence of osteolysis, and mechanical failure of a once annealed HXLPE out to 15-year follow-up.

This study had several limitations. First, there were patients who were lost to follow-up, meaning that there was the potential to miss outliers with high wear or implant failure. Second, there was no intraobserver and interobserver validation of radiographic wear measurements, but the authors believe the data are acceptable based on the Martell software. Repeatability of the measurements (on the basis of 5 measurements per patient) using Martell's technique was reported by Hui et al,21 with use of an intraclass correlation coefficient. Third, this was a nonconsecutive single-surgeon series that potentially could have biased the selection of patients. Fourth, the authors had no control cohort and were relying on historic reports. Fifth, an independent reviewer did not perform the review of the radiographs for osteolysis and implant fixation.

At an average follow-up of 12.2 years (maximum, 15 years), the annual linear wear rate was 0.018 mm (SD, 0.024 mm) and the annual volumetric wear rate was 11.1 mm3. These results compare favorably with those of reports of annealed and re-melted HXLPEs with 5 to 10 years of follow-up. The 5 studies in which re-melted HXLPE was implanted report a range of positive wear from 0.005 to 0.040 mm/y. One of the studies reported negative wear (−0.012 mm/y).25 The 6 studies, including the current one, with annealed highly cross-linked material had wear rates ranging from 0.001 to 0.037 mm/y. A multicenter study with the same once annealed HXLPE found 0.036 mm/y wear at 5 years and 0.031 mm/y wear at 9 years, representing a 72% and a 78% reduction, respectively, compared with N2 vacuum conventional polyethylene (Stryker Orthopaedics) (irradiated at 3 mrad). Not a single case had wear above 0.08 mm/y. A 9-year study of the clinical experience with the same once annealed material (Cross-fire) reported a wear rate of 0.037 mm/y at 9 years, and a recent study with 10-year follow-up of patients with an average age of 54 years reported a linear wear rate of 0.015 mm/y. The annual volumetric wear from 1 to 15 years (mean, 12.2 years) of 11.1 mm3 is comparable to the 14 mm3 from 1 to 10 years reported for a re-melted HXLPE. Osteolysis is reported to be a rare occurrence with annual volumetric wear below 80 mm3.

Wear debris–related osteolysis has not been found on routine radiographs out to 15-year follow-up. In addition, no osteolysis has been reported by the other 5 studies (range, 5–10 years) using annealed HXLPE. Two studies using re-melted HXLPEs have reported osteolysis rates of 5% and 10.6%.

No mechanical failures, either through fracture or disruption of the locking mechanism of annealed HXLPE, have been reported to date, in agreement with the current authors' experience. Although fractures have occurred with the re-melted materials, the reports have been isolated and related to the use of a thin polyethelene insert. Since its introduction, the concern regarding annealed HXLPE has been about the presence of free radicals and their long-term effect on clinical performance.

Preclinical bench or in vitro testing is a necessary part of total hip replacement bearing development. The ability of this tool to accurately predict clinical performance is variable and ultimately depends on the in vivo outcome. Several studies comparing once annealed with re-melted HXLPE subjected to accelerated artificial aging have found oxidation, decreased ultimate strength, and higher wear with the former. There is no debate that polyethylene, irradiated and once annealed, has free radicals. However, it shows excellent wear resistance, comparable with that of re-melted material, and superior strength. These artificial aging techniques do not always predict in vivo experience; they more closely parallel changes seen with components that are shelf aged. When materials were shelf aged for 4.3 to 9 years, the materials cross-linked and annealed showed significantly more oxidative damage than those cross-linked and re-melted. In an in vitro wear study of polyethylene subjected to 100 mrad of radiation in air and then aged for 30 years, samples triple packaged in air were compared with those shelf aged without packaging. The polyethylene cups aged in an air-containing package for 30 years had significantly lower wear in a wear simulator than samples shelf aged without packaging after 5 million cycles. This experiment is consistent with results from artificial aging techniques. The packaged material, although containing some oxygen at packaging, was shielded from further contact with oxygen. In contrast, the unpackaged was essentially bathed in oxygen for the duration of the aging process (similar to artificial aging) and showed a marked degradation and significantly increased wear compared with oxygen-deprived materials.

Of particular interest is a clinical study from early attempts (1970–1980s) at cross-linking with irradiation in air. Doses of 100 mrad in air (10 times the currently used doses) were used and followed for 25 years. This study showed no indication of increasing wear rates with time. In fact, just the opposite occurred, with no outliers having wear rates greater than 0.1 mm/y.

Retrievals of first-generation annealed components have shown evidence of in vivo oxidation, but the extent of oxidation is not uniform throughout the material. The rims of the annealed acetabular components routinely show oxidation, while the areas involving the locking mechanisms and bearing surfaces show minimal evidence, well below the critical level, of oxidative damage. Equally unexpected were changes in the first-generation re-melted components, which showed increasing evidence of oxidation (although still minimal) with increasing time in situ.

Until 1998, polyethylene cross-linking was an unintended consequence of gamma sterilization. The first attempt at enhanced cross-linking was Duration (Stryker Orthopaedics), which was irradiated at 3 mrad and annealed. With 14-year follow-up, its wear properties were an improvement over the predicate material but fell short of the dramatic improvement seen with the highly cross-linked materials introduced in 1998. Following the development of Crossfire, a sequentially annealed and irradiated HXLPE product named X3 (Stryker Orthopaedics) was manufactured to significantly reduce the presence of free radicals. At 7-year follow-up, using a 32-mm head, it has a reported linear wear rate of 0.018 mm/y and no osteolysis. The current authors have no clinical experience using 32-mm heads with Crossfire liners.

Thus far, one can conclude that laboratory testing and clinical experience agree regarding markedly improved wear resistance of HXLPE. No such agreement exists when accelerated aging techniques are compared with clinical data. Retrieval of HXLPE components yields some insight as to why this disconnect exists. The locking mechanism and bearing surface appear to be protected from in vivo oxidation either by shielding from oxygen-containing fluids and/or contact with the femoral head or by mechanisms yet to be described.

At this juncture, with solid clinical data now out to 10 to 15 years, both remelted and annealed materials show excellent clinical performance. Reduced mechanical properties in the former have not proved to be a major clinical problem and residual free radicals in the latter have not resulted in increased wear or mechanical issues.

Longer follow-up of well-designed studies is crucial to establishing whether free radical content in polyethylene components causes significant clinical problems. Once annealed HXLPE continues to have low wear without occurrence of osteolysis or mechanical failure out to 15-year follow-up.
 
References
  1. Harris WH. Conquest of a worldwide human disease: particle induced periprosthetic osteolysis. Clin Orthop Relat Res. 2004; 429:39–42. doi:10.1097/01.blo.0000149821.72760.39
    Dumbleton JH, Manley MT, Edidin AA. A literature review of the association between wear rate and osteolysis in total hip arthroplasty. J Arthroplasty. 2002; 17(5):649–661. doi:10.1054/arth.2002.33664
  2. Capello WN, D'Antonio JA, Ramakrishnan R, Naughton M. Continued improved wear with an annealed highly cross-linked polyethylene. Clin Orthop Relat Res. 2011; 469(3):825–830. doi:10.1007/s11999-010-1556-5
  3. Manning DW, Chiang PP, Martell JM, Galante JO, Harris WH. In vivo comparative wear study of traditional and highly cross-linked polyethylene in total hip arthroplasty. J Arthroplasty. 2005; 20(7):880–886. doi:10.1016/j.arth.2005.03.033
  4. McCalden RW, MacDonald SJ, Rorabeck CH, Bourne RB, Chess DG, Charron KD. Wear rate of highly cross-linked polyethylene in total hip arthroplasty: a randomized controlled trial. J Bone Joint Surg Am. 2009; 91(4):773–782. doi:10.2106/JBJS.H.00244
  5. Olyslaegers C, Defoort K, Simon JP, Vandenberghe L. Wear in conventional and highly cross-linked polyethylene cups: a 5-year follow-up study. J Arthroplasty. 2008; 23(4):489–494. doi:10.1016/j.arth.2007.02.013
  6. Thomas GE, Simpson DJ, Mehmood S, et al. The seven-year wear of highly cross linked polyethylene in total hip arthroplasty: a double blind, randomized controlled trial using radiostereometric analysis. J Bone Joint Surg Am. 2011; 93(8):716–722. doi:10.2106/JBJS.J.00287
  7. Tower SS, Currier JH, Currier BH, Lyford KA, Van Citters DW, Mayor MB. Rim cracking of the cross-linked longevity polyethylene acetabular liner after total hip arthroplasty. J Bone Joint Surg Am. 2007; 89(10):2212–2217. doi:10.2106/JBJS.F.00758
  8. Moore KD, Beck PR, Petersen DW, Cuckler JM, Lemons JE, Eberhardt AW. Early failure of a cross-linked polyethylene acetabular liner: a case report. J Bone Joint Surg Am. 2008; 90(11):2499–2504. doi:10.2106/JBJS.G.01304
  9. Wannomae KK, Christensen SD, Freiberg AA, Bhattacharyya S, Harris WH, Muratoglu OK. The effect of real-time aging on the oxidation and wear of highly cross-linked UHMWPE acetabular liners. Biomaterials. 2006; 27(9):1980–1987. doi:10.1016/j.biomaterials.2005.10.002
  10. Currier BH, Currier JH, Collier JP, Mayor MB, Citters DW. In vivo oxidation of highly cross-linked UHMWPE bearings. Paper presented at: 2012 Annual Meeting Orthopaedic Research Society. ; February 4–7, 2012. ; San Francisco, CA. .
  11. Reynolds SE, Malkani AL, Ramakrishnan R, Yakkanti MR. Wear analysis of first-generation highly cross-linked polyethylene in primary total hip arthroplasty: an average 9-year follow-up. J Arthroplasty. 2012; 27(6):1064–1068. doi:10.1016/j.arth.2012.01.006
  12. Ranawat CS, Palanisami DR, Landhey M. Ten-year wear data on a first-generation highly cross-linked polyethylene liner in young and active patients. Presented at: EFFORT 2012. ; May 23–25, 2012. ; Berlin, Germany. .
  13. Harris WH. Traumatic arthritis of the hip after dislocation and acetabular fractures: treatment by mold arthroplasty. An end-result study using a new method of result evaluation. J Bone Joint Surg Am. 1969; 51(4):737–755.
  14. Martell JM, Berdia S. Determination of polyethylene wear in total hip replacements with use of digital radiographs. J Bone Joint Surg Am. 1997; 79(11):1635–1641.
  15. Martell JM, Berkson E, Berger R, Jacobs J. Comparison of two- and three-dimensional computerized polyethylene wear analysis after total hip arthroplasty. J Bone Joint Surg Am. 2003; 85(6):1111–1117.
  16. DeLee JG, Charnley J. Radiological demarcation of cemented sockets in total hip replacement. Clin Orthop Relat Res. 1976; 121:20–32.
  17. Gruen TA, McNeice GM, Amstutz HC. “Modes of failure” of cemented stem-type femoral components: a radiographic analysis of loosening. Clin Orthop Relat Res. 1979; 141:17–27.
  18. Engh CA, Massin P, Suthers KE. Roentgenographic assessment of the biologic fixation of porous-surfaced femoral components. Clin Orthop Relat Res. 1990; 257:107–128.
  19. Hui AJ, McCalden RW, Martell JM, Mac-Donald SJ, Bourne RB, Rorabeck CH. Validation of two- and three-dimensional radiographic techniques for measuring polyethylene wear after total hip arthroplasty. J Bone Joint Surg Am. 2003; 85(3):505–511.
  20. Engh CA Jr, Hopper RH Jr, Huynh C, Ho H, Sritulanondha S, Engh CA Sr, . A prospective, randomized study of cross-linked and non–cross-linked polyethylene for total hip arthroplasty at 10-year follow-up. J Arthroplasty. 2012; 27(suppl 8):2–7. doi:10.1016/j.arth.2012.03.048
  21. Lee JH, Lee BW, Lee BJ, Kim SY. Midterm results of primary total hip arthroplasty using highly cross-linked polyethylene: minimum 7-year follow-up study. J Arthroplasty. 2011; 26(7):1014–1019. doi:10.1016/j.arth.2011.03.015
  22. Thomas GE, Simpson DJ, Mehmood S, et al. The seven-year wear of highly cross-linked polyethylene in total hip arthroplasty: a double-blind, randomized controlled trial using radiostereometric analysis. J Bone Joint Surg Am. 2011; 93(8):716–722. doi:10.2106/JBJS.J.00287
  23. Bragdon CR, Doerner M, Martell J, et al. The 2012 John Charnley Award: clinical multicenter studies of the wear performance of highly crosslinked remelted polyethylene in THA. Clin Orthop Relat Res. 2013; 471(2):393–402. doi:10.1007/s11999-012-2604-0
  24. Glyn-Jones S, Thomas GE, Garfjeld-Roberts P, et al. The John Charnley Award: highly crosslinked polyethylene in total hip arthroplasty decreases long-term wear. A double-blind randomized trial. Clin Orthop Relat Res. 2015; 473(2):432–438. doi:10.1007/s11999-014-3735-2
  25. D'Antonio JA, Capello WN, Mesko JW, Ramakrishnan R. Second-generation annealed highly crosslinked polyethylene has low wear at mean seven-year follow-up. Surg Technol Int. 2014; 25:219–226.
  26. Callary SA, Field JR, Campbell DG. Low wear of a second-generation highly cross-linked polyethylene liner: a 5-year radiostereometric analysis study. Clin Orthop Relat Res. 2013; 471(11):3596–3600. doi:10.1007/s11999-013-3188-z
  27. Oparaugo PC, Clarke IC, Malchau H, Herberts P. Correlation of wear debris-induced osteolysis and revision with volumetric wear-rates of polyethylene: a survey of 8 reports in the literature. Acta Orthop Scand. 2001; 72(1):22–28. doi:10.1080/000164701753606644
  28. Livingston BJ, Chmell MJ, Spector M, Poss R. Complications of total hip arthroplasty associated with the use of an acetabular component with a Hylamer liner. J Bone Joint Surg Am. 1997; 79(10):1529–1538.
  29. Sychterz CJ, Orishimo KF, Engh CA. Sterilization and polyethylene wear: clinical studies to support laboratory data. J Bone Joint Surg Am. 2004; 86(5):1017–1022.
  30. Kyomoto M, Ueno M, Kim SC, Oonishi H, Oonishi H. Wear of “100 Mrad” cross-linked polyethylene: effects of packaging after 30 years real-time shelf-aging. J Biomater Sci Polym Ed. 2007; 18(1):59–70. doi:10.1163/156856207779146150
  31. Oonishi H, Kuno M, Tsuji E, Fujisawa A. The optimum dose of gamma radiation: heavy doses to low wear polyethylene in total hip prostheses. J Mater Sci Material Med. 1997; 8(1):11–18. doi:10.1023/A:1018582027349
  32. Kurtz SM, Austin MS, Azzam K, et al. Mechanical properties, oxidation, and clinical performance of retrieved highly cross-linked Crossfire liners after intermediate term implantation. J Arthroplasty. 2010; 25(4):614–623. doi:10.1016/j.arth.2009.04.022
  33. Reinitz SD, Currier BH, Franklin KJ, Tibbo ME, Van Citters DW, Collier JP. Early indication of oxidative degradation in retrieved annealed UHMWPE bearings resembles gamma sterilized materials. Paper presented at: 2012 Annual Meeting Orthopaedic Research Society. ; February 4–7, 2012. ; San Francisco, CA. .
  34. Muratoglu OK, Wannomae KK, Rowell SL, Micheli BR, Malchau H. Ex vivo stability loss of irradiated and melted ultra-high molecular weight polyethylene. J Bone Joint Surg Am. 2010; 92(17):2809–2816. doi:10.2106/JBJS.I.01017
  35. Grimm B, Tonino A, Heyligers IC. Fourteen years long-term performance of crosslinked PE: sustained wear reduction and less osteolysis? Paper presented at: 2012 Annual Meeting Orthopaedic Research Society. ; February 4–7, 2012. ; San Francisco, CA.
 